    Dear all,

    I’ve just trying to get LODs for som analyses where I don’t have an IS value, so, the semiquantitative approach is perfect for me. In the iolite user manual there is a good description for how iolite calculates LOD but only for the Internal Standardization Method. Where could I found some explanation about how iolite calculates LOD in case I use the semiquantitative method on Trace Elements DRS. And also, if possible, get also explanation about how exactly the Traces Element DRS calculates the semiquantitative approach concentrations…

    Thanks a lot

    Regards

    Manuel J. Román – Instituto Andaluz de Ciencias de la Tierra – Granada / Spain

Viewing 1 replies (of 1 total)
  • Author
    Replies
  • #9429

    Hi all,

    This question was answered off forum by Joe. Here is the answer in case someone else needs this info:

    The semi-quantitative LODs use the semi-quantitative sensitivity, i.e., the count rate produced by X in the reference material / the concentration of X in the reference material. Otherwise, the equations are the same.